If the Pinpoint ORP calibration solution is like any of the other premade ORP solutions, then it is not very accurate. Anyway you'll have to manually adjust the ORP offset if you calibrate with a reference we don't support. Here is how you can do it:

Put the ORP probe in the 400 solution.
With the controller in the run display note the ORP reading.
Now go to the hidden manual calibration menu. To do so go to the self test display (second screen on AC2). Press the up button twice, the down button twice, and then the select button.
Select ORP manual cal.
Adjust the ORP offset up or down depending on the reading in the run display. For example say the ORP probe reads 390 instead of 400, then the ORP offset needs to be increase by 10. Leave the ORP scale unchanged since this is a single point calibration.

 I thought this was interesting so I thought I'd share . This is how to calibrate orp on the neptune with only 1 test fluid

Edited by bugzme
Jeff
125 tank
50 gallon sump
T-5 lighting
Rum drinker, Carbon User
I KNOW ROCKS THAT ARE YOUNGER THEN ME!! I AM A Realist! I write what I think!!
Back to Top
 Post Reply Post Reply
  Share Topic   

Forum Jump Forum Permissions View Drop Down

Forum Software by Web Wiz Forums® version 12.03
Copyright ©2001-2018 Web Wiz Ltd.

This page was generated in 0.125 seconds.